Originally Posted by cryo-tek
Good morning,
Pythsgoraa is correct, the Polycold units use Zerol 150 in most of there cascades. Telemark Cryogenics uses POE theres.
The job of an oil separate or a phase separator is to separate the liquids from the gas.
The two have the same function. Oil in a hot discharge line is hard to trap. If the discharge line has been cooled the oil is easier to trap. It gets easer if there is a high boiler refrigerant that will mix with the oil.(R-11,113,114,123)
The oil management in an autocascade is easer than in a mulit compressor cascade. In the autocascade, the oil must pass through the phase separators before it reaches the low temp stages. The oil is separated out of the discharge gas along with the condensed refrigerant. The bulk of the oil is removed from the gas stream along with the highest boiling refrigerants in the first separator. (The high boilers need to be miscible with the oil used) The oil mist that passes through the separator is cooled in the cascade condenser along with the refrigerants. The condensed refrigerants and oil are than separated from the gas in the second separator.
I love this but I have to do some real work.
I’m still new at this thread thing, is this the right place to ask a question or do I need to start a new thread? If this is the wrong place please tell me and I will move it
My problem: A new unit that was built has lower suction pressures that the standard units. It also performs better. A new person charged the unit. The rest should be the same but it’s not. What could affect suction pressure?
Captubes
Compressor pumping speed
?
?